Appendix A.Specifications

Available Ranges¹ 

100, 150, 200, 300, 500, 600, 1000, 1500, 2000 kip

Rated Capacities2

100 kN to 10,000 kN

Accuracy3

±0.5% F.S.

Output

1.5 mV/V to 2.5 mV/V @ F.S.

Linearity

0.5% F.S.

Resolution4 

0.025% F.S.

Repeatability5

0.1% F.S.

Temperature Effect

0.02% F.S./°C

Temperature Range

−20 °C to +80 °C
0 °F to 110 °F

Overrange6

150% F.S.

Input Resistance

350, 700, 1050, or 1400

Output Resistance

350, 700, 1050, or 1400

Excitation Voltage

2 to 15 VAC or DC

Maximum Excitation Voltage

30 V

Cable Type7

Three twisted pair (six conductor) 22 AWG
Foil shield, PVC jacket, nominal OD = 9.5 mm (0.375")

table 3: Model 3000 Load Cell Specifications

Notes:

1 Other ranges available on request.

2 Other capacities available on request. Calibrations that exceed geokon’s NIST traceable capacity of approximately 10,675 kN are subcontracted to an accredited testing laboratory.

3 The accuracy of the testing machine used to calibrate the cells is ±¼% F.S. traceable to NIST. The system accuracy depends on end loading conditions as described in the text of this manual. If field conditions are well duplicated during actual calibration the accuracy should be within ±5 F.S. Failure to duplicate conditions can cause calibration variations of ±20% in extreme cases.

4 Minimum, depends on the readout instrument and technique.

5 Repeatability under the same loading conditions. This does not take into account hysteresis and any changes in the loading conditions.

6 With no calibration shift. The maximum overrange without failure is 200% F.S. for aluminum cells, 300% to 400% F.S. for steel cells, but both the no-load zero and the calibration may change if the load exceeds 150% F.S.

7 Other cable types available.