Appendix B.Wiring And Connector Pinouts
The GK-502 is set up for remote sense of the excitation voltage at the load cell, which requires a six-conductor cable (three twisted pairs). If the load cell is not set up for remote sense and only has four conductors, then it will be necessary to modify the cable plug so that pins J and B and pins K and C are shorted together, otherwise the load cell cannot be read.
The P3 readout box will not read remote sense therefore load cells that have been wired for remote sense must be rewired at the connector for non-remote sense.
B.1Wiring Diagram with P3 Readout
6:
Figure 6: P3 Wiring
B.2Wiring Diagram with Remote Sense (GK-502 Readout)
7:
Figure 7: Remote Sense (GK-502) Wiring
B.3Wiring for use with GK 502 Readout Box
Bulkhead (10−pin) |
Amp (6−pin) |
Circuit Label |
Description |
Internal Load Cell Wiring |
Purple Cable |
A |
C |
S- |
Bridge Output - |
White |
White's Black |
B |
B |
P+ |
Bridge Excitation + |
Red |
Red |
C |
A |
P- |
Bridge Excitation - |
Black¹ |
Red's Black |
D |
D |
S+ |
Bridge Output + |
Green¹ |
White |
E |
|
NC |
No Connection |
|
NC |
F |
E |
G |
Ground for shield |
|
Bare Drain Wires |
G |
|
NC |
No Connection |
|
NC |
H |
|
NC |
No Connection |
|
NC |
J |
|
RS+ |
Remote Sense + |
Red² |
Green |
K |
|
RS- |
Remote Sense - |
Black² |
Green's Black |
table 4: GK-502 Wiring
Notes:
¹ Green and black wires switched on geokon load cells prior to serial number 1190.
² Non-remote sense is optional and must be specified at the time of ordering.
Bulkhead (10−Pin) |
P3 Label |
Description |
Purple Cable |
B |
P+ |
Bridge Excitation + |
Red |
C |
P- |
Bridge Excitation - |
Red's Black |
A |
S- |
Bridge Output - |
White's Black |
D |
S+ |
Bridge Output + |
White |
|
D120 |
No Connection |
NC |
|
D350 |
No Connection |
NC |
|
D1K |
No Connection |
NC |
F |
Shield |
Shield Connection |
Bare Drain Wires |
table 5: P3 Wiring