4.Data Reduction

4.1Displacement Calculation

The basic units utilized by geokon for measurement and reduction of data from vibrating wire displacement transducers are digits. Calculation of digits is based on the following equation:

Digits_Calculation.png

 

equation 1: Digits Calculation

To convert digits to displacement the following equation applies:  

Duncorrected = (R1 R0) x G x F

equation 2: Displacement Calculation

Where:

R1 is the current reading.

R0 is the initial reading, usually obtained at installation.

G is the gauge factor, usually millimeters or inches per digit.

F is an optional engineering units conversion factor (see the table below).

From

To

Inches

Feet

Millimeters

Centimeters

Meters

Inches

1

12

0.03937

0.3937

39.37

Feet

0.0833

1

0.003281

0.03281

3.281

Millimeters

25.4

304.8

1

10

1000

Centimeters

2.54

30.48

0.10

1

100

Meters

0.0254

0.3048

0.001

0.01

1

table 2: Engineering Units Conversion Multipliers

For example, if the initial reading R0 is 6783 digits, the current reading, R1, is 7228. The calibration factor, G, is 0.011906 mm/digit. The displacement change is:

+5.3 mm = (7228 6783) x 0.011906

equation 3: Displacement Change

Note that increasing readings (digits) indicate increasing extension.

4.2Temperature Correction

geokon’s vibrating wire displacement transducers have a small coefficient of thermal expansion. Correction may not be necessary in most cases. However, to achieve maximum accuracy, there are corrections that you can apply.

Use the following equation to provide thermal correction of the instrument:

Dcorrected = G(R1  R0) + K(T1  T0) 

equation 4: Thermally-Corrected Displacement Calculation

Where:

R1 is the current reading

R0 is the initial reading

G is the linear gauge factor

T1 is the current temperature

T0 is the initial temperature

K is the thermal coefficient (see Equation 5)

 

Tests have determined that the thermal coefficient, K, changes with the position of the transducer shaft. The first step in the temperature correction process is determination of the proper thermal coefficient based on the following equation:

K = ((R1 x M) + B) x G

equation 5: Thermal Coefficient Calculation

Where:

R1 is the current reading

M is the multiplier

B is the constant

G is the linear gauge factor from the supplied calibration sheet.


Model:                        

Multiplier (M):

Constant (B):

4450-3 mm (0.125")

0.000520

3.567

4450-12 mm (0.5")

0.000375

1.08

4450-25 mm (1")

0.000369

0.572

4450-50 mm (2")

0.000376

0.328

4450-100 mm (4")

0.000398

0.0864

4450-150 mm (6")

0.000384

-0.3482

4450-200 mm (8")

0.000396

-0.4428

4450-230 mm (9")

0.000403

-0.5016

4450-300 mm (12")

0.000424

-0.6778

table 3: Thermal Coefficient Calculation Constants

Consider the following example using a Model 4450-200 mm displacement transducer:

R0 = 4250 digits

R1 = 5875 digits

T0 = 10 °C

T1 = 20 °C

G = 0.06152 mm/digit

K = (((5875 x 0.000369) + 0.572) x 0.06152) = 0.0168

Dcorrected = ((R1 – R0) x G) + ((T1 – T0) x K)

Dcorrected = ((5875 – 4250) x 0.006152) + ((20 – 10) x 0.0168)

Dcorrected = 9.997 + 0.168

Dcorrected = +10.165 mm


The temperature coefficient of the mass or member to which the displacement transducer is attached should also be taken into account. Use the temperature coefficient of the mass or member, combined with the changes in temperature from initial to current readings, to determine thermal effects of the mass or member.

4.3Environmental Factors

Because the purpose of using a displacement transducer is to monitor site conditions, factors which may affect these conditions should always be observed and recorded. Seemingly minor effects may have a real influence on the behavior of the structure being monitored and may give an early indication of potential problems. Some of these factors include, but are not limited to: blasting, rainfall, tidal levels, excavation and fill levels and sequences, traffic, temperature and barometric changes, changes in personnel, nearby construction activities, seasonal changes, etc.12:    

image14.png

Figure 12: Typical Calibration Sheet